9.8?! I've never seen tapwater that high...better double check that with another test kit or at least clean your vials and test again. If you're using electrodes better recalibrate with at least a two point cal (7.0 and 10.0)and test again. If it's still off, clean the electrode using electrode cleaning sol'n, make sure there are no bubbles in the electrode and that salts haven't corroded the connections/internals and refresh the calibration.
Totally pure water should be pH neutral, or 7.0. I've found most test kits to be off a little, but shouldn't be off by more than 0.1 or 0.2 units (electrodes should be much closer if cal'ed properly). Remember, the pH scale is logarithmic (a small difference in pH means a big diff in [H+]. So, make sure your vials are clean and your reagents are not fouled. If you have a TDS meter on your RO/DI unit it should read as close to 0 as possible - no more than a very few ppm on the output water. You may need to replace your media if it's reading high. Don't have one? You can pick up a handheld unit at most decent LFS or online for about 50-75 bucks.