I agree with Scoots...as you add drops of the solution from the syringe you move up the Ca scale...if you are pouring the whole syringe into your test tube with no color change your Ca reading is too high, not zero.
Also just for sh&% and giggles are you swirling the test tube after each drop, I believe this is an important step as it help to aggitate the sample and drive the reaction.
Something is seriously off if your Calcium is over 500 and your Alk is only 3.8dkh. Even the worst salt mixes shouldn't yield this result.... and Tropic Marin is far from the worst. To raise your Alk, I assume you are only adding part 1 of the B-Ionic. Adding both parts at this point is counterproductive and you will quickly start precipitating calcium from solution.
since you were misinterpreting the Ca. test results, could you be doing the same for the Alk?
Yes, I am only using part 1. However, I am not misinterpreting the result of the Ca test. It is either pink or blue. When I run the test it stays pink.
Staying pink means the calcium level hasn't yet been reached via titration. It doesn't mean the level is 0. I think that's what Randy (prattreef) was referring to when he said "misinterpreting".
Dean is correct. If the solution hasn't turned color and the syringe is empty your CA level is above 500 ( the 1st line of the table-0ml= 500ppm) Similarly, an alkalinity reading of 3.8 dKH would leave the syringe full of .76 ml when the final reaction occurred ( blue turns pink). Is this the case? RD
I don't have the table in front of me for the Ca, but I thought the conc. decreased as more solution from the syringe was used. But if you say so I believe that the conc. is >500.
if that is the case it is very strange. I've never encountered such a huge disparity particularly in a new system. I'd mix up a small fresh batch of salt, let it sit over night and test it. The 2 most likely culprits are a bad batch of salt or a bad Alk. kit.
If I am not mistaken you move up the scale on both tests as syringe empties. I agree with Deanos and Randy. Perhaps you an try the more sensitive test scenario. I believe the test kits suggests you do this my cutting the sample water in half. Perhaps that can give you a more approptiate reading. (not sure if that will work)
Now on the point with dosing b-ionic (which is what I believe you mentioned you are dosing with) are you adding both parts or only one. Is it wise to dose with one and not the other, is there a relationship between the 2 parts that could be causing these readings. (EXPERTS please chime in, I have no answers. Just posting quesions to help out.)